Wrong voltage dysplay

Post any questions you may have about our current range of oscilloscopes
Post Reply
vadim_klimov
Newbie
Posts: 0
Joined: Wed Nov 06, 2019 7:39 am

Wrong voltage dysplay

Post by vadim_klimov » Thu Nov 07, 2019 7:55 am

I 'm new here and have little knowledge of thing. I have an pulse generator and would like to measure the signals with PicoScope. I have set the voltage of signal to 1V and can see that in analog oscilloscope as well. But picoScope shows me always 2V. What could be the cause?

Martyn
Site Admin
Site Admin
Posts: 3724
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Wrong voltage dysplay

Post by Martyn » Thu Nov 07, 2019 11:25 am

What is the output impedance of the pulse generator, and what is the input impedance of the analog oscilloscope ?
Martyn
Technical Support Manager

vadim_klimov
Newbie
Posts: 0
Joined: Wed Nov 06, 2019 7:39 am

Re: Wrong voltage dysplay

Post by vadim_klimov » Thu Nov 07, 2019 12:39 pm

Output impedance of the pulse generator is 50 ohms, and the input impedance of the analog oscilloscope is 50 ohms too.

Martyn
Site Admin
Site Admin
Posts: 3724
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Wrong voltage dysplay

Post by Martyn » Thu Nov 07, 2019 1:05 pm

As our scopes have a 1 MOhm input impedance you would read a 1V output from a 50 Ohm pulse generator at around 2V.
Martyn
Technical Support Manager

vadim_klimov
Newbie
Posts: 0
Joined: Wed Nov 06, 2019 7:39 am

Re: Wrong voltage dysplay

Post by vadim_klimov » Thu Nov 07, 2019 1:22 pm

Oh, sorry, i founded it. I must select 50 ohms dc